Clinical Using of 10FFF
I’ve seen several clinics routinely using 10 FFF, even for brain SRS. How frequently does your clinic use 10 FFF, especially for brain cases? While it seems clinically acceptable, I generally prefer 6 FFF to preserve a sharper penumbra and spare normal brain tissue. I’d love to hear your perspectives.
10
Upvotes
15
u/KrimsonKing Therapy Physicist 27d ago
10FFF for liver SBRT breath hold. Maybe other non-lung SBRT. Not much else.
11
u/IGRT_Guy Therapy Physicist 27d ago
The way I've always talked this through with fellow physicists, residents, and physicians is to look at the control points for your delivery.
At 20 Gy SRS, 10FFF generally doesn't hit max dose rate. 6FFF can, so the gantry slows down, but not by much. You're talking seconds per arc, not 50% slower. Once you get into the 25+ Gy range, that's where 10FFF becomes a real time saver.
14
u/s32bangdort 27d ago
I would not use 10fff for brain SRS exactly because of the penumbra.
5
u/StopTheMineshaftGap 27d ago
If you ever begin functional SRS or breath hold SBRT, you will love it.
The penumbra differences are not usually clinically relevant and washout with good planning.
There are additional theoretical BED advantages as well.
11
u/ericvt Therapy Physicist 27d ago edited 27d ago
The penumbra difference for 10FFF compared to 6FFF is largely negligible for intracranial applications. Planning technique / experience is a much more impactful quality differentiator in terms of driving dose compactness. Even for trigeminal neuralgia treatment there are published outcome studies where 10FFF is used to excellent effect (NCT02708810), as well as ~130 Gy radiosurgical thalamotomy (DOI: 10.1016/j.ijrobp.2020.07.1509).
In my experience, the only place where 6FFF is appreciably better than 10FFF is in a specific subset of lung SBRT scenarios where you are treating a very small tumor within an emphysemic patient (or who otherwise has atypically low density lung tissue) and buildup effects at the lung-tumor interface become more important. Even then you have to carefully weigh the tradeoff between dosimetric benefit and efficiency.
6
u/Sea_Toe6263 27d ago
We've stopped 10FFF a few years ago because it didn't save us much time on the machine. Didn't seem worth the it for the issues you mentioned.
5
u/No-Occasion251 27d ago
Excited to try out 8fff on the new machine. Faster MLC and gantry might lead to dose rate limiting
2
u/_Shmall_ Therapy Physicist 26d ago
I like 6fff in brain cases for the sharp penumbra. Then mix of 10fff and 6fff for everything else. Try a 12 Guy per fraction lung SBRT with 6 and 10 FFF at the machine. Time savings of almost half with 10 FFF.
2
u/PA_Med_Physicist Therapy Physicist 24d ago
2400 MU/min seems quite necessary for gated SBRT.
1
u/BeachAggravating7870 23d ago
The only case we used it for is a single shoot, spine 8gy
1
u/Vernost Therapy Physicist 22d ago
Recently had an interesting case that was a gated lung SBRT 34 Gy x 1 fx. In this case the dose rate improvements outweighed the dosimetric difference with beam quality. Agree with one of the other comments that oftentimes plan quality is much more dependent on dosi skill and optimization than just beam energy.