Clinical Using of 10FFF

I’ve seen several clinics routinely using 10 FFF, even for brain SRS. How frequently does your clinic use 10 FFF, especially for brain cases? While it seems clinically acceptable, I generally prefer 6 FFF to preserve a sharper penumbra and spare normal brain tissue. I’d love to hear your perspectives.

10 Upvotes

23 comments sorted by

15

u/KrimsonKing Therapy Physicist 27d ago

10FFF for liver SBRT breath hold. Maybe other non-lung SBRT. Not much else.

8

u/6mvphotons 27d ago

Yep, liver and prostate, and sometimes a para-aortic node SBRT. We'd never use it for a brain.

11

u/IGRT_Guy Therapy Physicist 27d ago

The way I've always talked this through with fellow physicists, residents, and physicians is to look at the control points for your delivery.
At 20 Gy SRS, 10FFF generally doesn't hit max dose rate. 6FFF can, so the gantry slows down, but not by much. You're talking seconds per arc, not 50% slower. Once you get into the 25+ Gy range, that's where 10FFF becomes a real time saver.

6

u/womerah Therapy Resident 27d ago

Is 10FFF even that much of a time saver in the 25Gy+ range vs 6FFF? For the whole appointment including setup, image matching, shifts, verification, etc - I feel it must be only a minor saving.

10FF vs 10FFF is where the time savings are, no?

2

u/IGRT_Guy Therapy Physicist 27d ago

I was thinking of mv beam on to mv beam
Off time. Once you start that you are committed, ie breath hold, super full bladder, painful Mets

2

u/womerah Therapy Resident 26d ago

Ah gotcha, a second or two of less breath hold might be the difference between being a DIBH or free breather

14

u/s32bangdort 27d ago

I would not use 10fff for brain SRS exactly because of the penumbra.

5

u/StopTheMineshaftGap 27d ago

If you ever begin functional SRS or breath hold SBRT, you will love it.

The penumbra differences are not usually clinically relevant and washout with good planning.

There are additional theoretical BED advantages as well.

2

u/ClinicFraggle 27d ago edited 27d ago

BED advantages over 6 FFF? Can you elaborate or add any reference, please?

I think the question is if the (probably small) advantages of 10FFF for a few treatments warrant the extra resources and time required for commissioning and periodic QC (output, profiles, Winston-Lutz, etc)

There may be some difference in usage between 10 FFF in Varian and Elekta, because one is a little less energetic than the other (perhaps it could make more sense in Varian)

1

u/StopTheMineshaftGap 27d ago

Check out the old trigeminal BED literature based on Gamma Knife source age.

11

u/ericvt Therapy Physicist 27d ago edited 27d ago

The penumbra difference for 10FFF compared to 6FFF is largely negligible for intracranial applications. Planning technique / experience is a much more impactful quality differentiator in terms of driving dose compactness. Even for trigeminal neuralgia treatment there are published outcome studies where 10FFF is used to excellent effect (NCT02708810), as well as ~130 Gy radiosurgical thalamotomy (DOI: 10.1016/j.ijrobp.2020.07.1509).

In my experience, the only place where 6FFF is appreciably better than 10FFF is in a specific subset of lung SBRT scenarios where you are treating a very small tumor within an emphysemic patient (or who otherwise has atypically low density lung tissue) and buildup effects at the lung-tumor interface become more important. Even then you have to carefully weigh the tradeoff between dosimetric benefit and efficiency.

3

u/Conscious_Platypus10 26d ago

This is the answer

2

u/s32bangdort 26d ago

In the era of multi-met SRS now taking over the role that whole brain used to, I disagree. The more we can spare the normal brain tissue the better. Sure in single target like trigem it’s probably fine but absolutely should be considered for multi-met SRS. I wouldn’t just hand wave it away with planner inconsistency. I would expect more out of my planners.

8

u/ericvt Therapy Physicist 26d ago

I'm not saying there isn't a benefit, I am saying that the benefit is so modest that it is a secondary or tertiary effect and is further diluted by competing effects like differences in planner approaches.

The most upvoted comment in this thread implies that 10FFF shouldn't be used in SRS because of the penumbra, which is just not a defensible statement. We're talking ~0.3 mm difference in 20% - 80% penumbral width and a few percent relative difference in intermediate/low dose spill volumes, and the difference reduces as target count goes up. Yet there are still planners out there eyeballing what collimator angle they should use in multimet plans instead of using software to algorithmically solve it, for example.

1

u/Beam_Hardener 24d ago

Any software recommendations/solutions for collimator angle optimization?

5

u/ericvt Therapy Physicist 23d ago

If you are in an Eclipse environment then Hyperarc is the commercial option, but it is a relatively straightforward thing to script out using ESAPI. There are several papers that have been published over the years about optimal collimator angles in multi-met SRS. The route we took with ESAPI is:

  • Compute the recommended isocenter position at the centroid midpoint of the bounding box of all PTV centroids
  • Project each PTV mesh onto the BEV plane at SAD (accounting for divergence), for every candidate gantry angle sampled at 5 degree intervals across each arc (gantry angles come from templated sets of arc arrangements).
  • Sweep candidate collimator angles in 5 degree increments and quantify inter-PTV unshielded area at each collimator angle by computing the union area of gaps between PTV projections in the BEV, as well as mean enclosing jaw-defined field area
  • Select the optimal collimator angle per arc by ranking candidates: minimum unshielded area, then minimum mean field area, then proximity to cardinal angles, etc.
  • Fit jaw positions per arc to the tightest rectangle enclosing all PTV projections across all gantry samples at the selected collimator angle, with a small margin added on each side (starting point for jaw tracking).
  • Apply the recommended isocenter, collimator angles, and jaw positions to the open plan

So essentially the code iterates through all collimator angles at each gantry position (both in 5 degree increments) and finds what collimator angle minimizes the sharing of leaf pairs between any lesions. This is part of our larger SRS autoplanning script, but the collimator angle optimization by itself would be about ~400 lines of code.

1

u/Logical-Pattern8065 22d ago

A lot of planners only look at the GI and, yes, usually better for 6xFFF. But with zero or near zero margins, a bit of penumbra with 10xFFF might be a win for tumor control. Biology rules in the end, not physics.

6

u/Sea_Toe6263 27d ago

We've stopped 10FFF a few years ago because it didn't save us much time on the machine. Didn't seem worth the it for the issues you mentioned.

5

u/No-Occasion251 27d ago

Excited to try out 8fff on the new machine. Faster MLC and gantry might lead to dose rate limiting

2

u/_Shmall_ Therapy Physicist 26d ago

I like 6fff in brain cases for the sharp penumbra. Then mix of 10fff and 6fff for everything else. Try a 12 Guy per fraction lung SBRT with 6 and 10 FFF at the machine. Time savings of almost half with 10 FFF.

2

u/PA_Med_Physicist Therapy Physicist 24d ago

2400 MU/min seems quite necessary for gated SBRT.

1

u/BeachAggravating7870 23d ago

The only case we used it for is a single shoot, spine 8gy

1

u/Vernost Therapy Physicist 22d ago

Recently had an interesting case that was a gated lung SBRT 34 Gy x 1 fx. In this case the dose rate improvements outweighed the dosimetric difference with beam quality. Agree with one of the other comments that oftentimes plan quality is much more dependent on dosi skill and optimization than just beam energy.