UT1 Studying question
I’m studying for my UT1 and was wondering if someone could clarify the answer to a question for me.
In the ASTM UT Q&A third edition, question 87 ask:
“A separate time baseline imposed on the viewing display of some ultrasonic testing instruments that permits measurement of distances is often referred to as:
a. an initial pulse.
b. a time/distance line.
c. an electronic gate.
d. a sweep line.”
To me the answer that makes the most sense is B but the book list the correct answer as C, could anyone help me understand why the answer is C?
Thanks!
4
u/Business_Door4860 3d ago
This is a question where you need to understand each answer and why it isn't a the correct one, but the key word is "separate".
2
u/OhAces 3d ago
It is kindof poorly worded, but they mean gate for sure. The baseline is already the time/distance line. The gate is purely for measurements. For process of elimination, you know for sure its not the IP or a sweep line, that would be vertical, you already have the t/d line, that leaves the gate.
2
u/VirtualLightning 3d ago
"Permits measurement of distance" ... the gate will provide measurements within its area determined by the delay, length and if its dynamic its movement. This rules out the amplitudes appearing before or after, "outside" of the gate.
2
u/Prestigious-Band7502 3d ago
The confusion is understandable! The sweep is the ruler, while the electronic gate selects the range being monitored.
2
u/Novel_Company_5867 2d ago
It's written from the perspective of analogue instrumentation, hence the "some ultrasonic testing instruments" part. The last gasp on the analogue front was to superimpose a horizontal line on the A-scan CRT display to represent a gate.
They seriously have to revise their questions. Analogue instruments were on the way out 40 years ago.